- Posts: 9
- Joined: Sun Aug 10, 2014 11:31 am
FID Not responding
Discussions about GC and other "gas phase" separation techniques.
6 posts
Page 1 of 1
Agilent 6890 Series GC with FID and TCD showing Front detector not found (FID). What can be the possible cause. Please help
-
- Posts: 5329
- Joined: Thu Oct 13, 2005 2:29 pm
- Location: Maun Botswana
FID not connected ?
Peter Apps
-
- Posts: 318
- Joined: Sun Jan 01, 2012 10:21 am
- Location: Turkiye
Temperature sensor is open circuit ( broken )
-
- Posts: 9
- Joined: Sun Aug 10, 2014 11:31 am
uzman wrote:
Temperature sensor is open circuit ( broken )
This may be the cause. In the display panel in front of the GC it is showing Temp Not installed. If this is the reason is it required to buy only the heator sensor assembly or any other items?Kindly help
-
- Posts: 318
- Joined: Sun Jan 01, 2012 10:21 am
- Location: Turkiye
Temperature sensor and the heater assy is sold together , and it is better ( and easier ) to change both.
Sensor is a tiny and fragile part so care is needed during installation.
Sensor is a tiny and fragile part so care is needed during installation.
-
- Posts: 3
- Joined: Sat Jun 13, 2020 3:52 pm
bijish wrote:
Agilent 6890 Series GC with FID and TCD showing Front detector not found (FID). What can be the possible cause. Please help
Did you solve the problem. I encountered the same problems. And we have same issue here.
When I was doing an experiment. Suddenly, The thermal system had a fault. All the temperature was off. So I waited for everything cooled down. Shutdown the GC system completely. After that, I restarted the GC. I increased the FID’s temperature to 200C. Everything was good. As I increasing the temperature to 400C. FID’s temperature cannot stable at 400C. There was much oscillation here. Then, the front panel of FID showed “ Temp not installed”.
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science