Problem: Operator changed the column on 5/8. (Column is DB-Wax, 15m x 0.25mm x 0.25 um) After changing the column, the first injection of a sample after the instument has been sitting idle for a while, shows reduced sensitivity of ALL peaks and a slight RT shift to higher time. A second injection, from the same vial, directly after will shows correct response and RT's. All other vials in same sequence run are also okay. The oven temp at idle conditions is 130 deg C. Method is Linear Temp Program with final temp of 230 deg C held for two minutes to bake out column.
Troubleshooting Steps Taken: 1. Syringe was changed and problem seemed to stop, only to return a day or so later. 2. Check of column nuts and underside of inlet revealed that the ferrule was not crimped down against the gold seal. Changed gold seal, inlet liner, new ferrule and recut column before replacing in inlet. Changed septum and tightened nut. Checked and tightened detector nut. Okay for a day or two 3. Problem reappeared. Detector side of column was removed and ferrule showed proper crimping and correct length inserted into FID detector. Replaced and leak checked. No detectable leaks were found 4. 5/23/12 - Service Engineer came in.........Blank inlet nuts and purge gas caps were placed on instrument by service engineer to isolate inlet for pressure check. All pressure checks and flow checks were normal.
In 30 years of doing GC, I've never seen this happen. Only the 1st sample of a sequence run is affected, so it's not related to anything that would affect every sample (dirty FID, clogged purge line, etc.). It's like a leak that only shows up when the instrument sits idle and then once everything is warmed up and "expanded" after the first run, it seals up and runs everything normally.
Has anyone else ever seen this happen before? If so, what caused it? Even our service engineer is stumped by this one. I'm fresh out of ideas.......
Thanks