Dear all,

I was hoping if anyone can give me some help on this, and thank you in advance!

We have a Shimadzu GC-2010+, and installed an Agilent GC Gaspro 60 m x 0.320 mm. We want to detect methane and ethane in Ar. We use a gas syringe to draw samples from a gas bag and inject into the GC manually. We tried different size syringe (0.1 mL, 1 mL, 5 mL)

We see the methane and ethane peaks, but they always split into 2 peaks. Sometime they split equally and sometimes they split into a big and a small. However, the split ratio between those two peaks are always very similar.

The Method
Image
https://i.postimg.cc/fRTG8xbQ/Weixin-Im ... 184605.jpg
Image
https://i.postimg.cc/nhPPTg5q/Weixin-Im ... 184614.jpg

Problem Pictures
Image
https://i.postimg.cc/W1S4JHLg/Picture1.jpg
Image
https://i.postimg.cc/QtRYs7sC/Weixin-Im ... 184221.jpg
Image
https://i.postimg.cc/bYk3ksp0/Weixin-Im ... 184224.jpg

Does anyone know what happened? Why would the peak split in such a way?

Thank you!!
It looks to me like you have plenty of signal, if I'm you I would try a split injection. You can adjust the split ratio to get the sensitivity you need. I bet your peak shapes improve dramatically if you split the injection.
claserj,

I'm with rb6banjo. YOu can certainly stand to go to a split injection at a minimum of something like 10:1 based on what I see. The other thing I notice is your column flow is really low. 0.32's run best around 2-4 mL's min and you are supplying less than 1. Try taking that up to around 2 or 3 and see what happens.

Best regards,
Dear rb6banjo and AICMM,

Thank you for your help. We injected the samples in split mode 1:10 and the peaks no longer split. We also adjusted (reduced) the flow rate, and now the peaks have less tailing. Thank you!

Image
https://i.postimg.cc/pXggRqr6/20230723115028.png

Image
https://i.postimg.cc/3N2bRHdf/20230723115028-1.png

Image
https://i.postimg.cc/YCvYfzxM/20230723115028-2.png

Image
https://i.postimg.cc/Qx2xwg3X/20230723115028-3.png
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry