May any one tell me that why hump is coming g in the chromatogram of GCMS with J&W DB-5MS column. Column is new and conditioning g has been preformed?
Is this a rise in baseline at the end of the chromatogram on a temperature programmed run or is it in the middle of the chromatogram?

Also what is the dominate mass present in the spectra of the hump you are seeing?
The past is there to guide us into the future, not to dwell in.
Hello Saba,

I agreed with James, it's depending on where are you get the hump. Normally the column bleed would occur with raising temperature, you could have a look on the oven temperature.
Then, the new o-rings and septum is an potential source of the hump as well.
Basically, mass on 207 (major), 281, and 73 are shown for the column bleed.

Hope this would be useful on your case.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry