I'm retired now, and our 5890 units were hated by my last supervisor. But they worked great with the 7673 autosampler. Let's assume that the 9-pin remote cables are all in place, and that the Configuration Editor has been updated to accept the autosampler. Which control box are you using for the autosampler?

You can program in a post run, or the system will inject again once all the temperature parameters are reached and held.
Hi. Thank you for your response and your time.
So - the issue was the remote cable from GC was not connected to the controller. Now everything is working.
We now need to work on the method for short-chain fatty acids. We configured the GC with the following parameters:
Oven: 125C - 20 minutes
Inlet: 175 C
Detector: 180 C
Flow A/Nitrogen - 40mL/min
Flow B/Helium - 6

However, I am unfamiliar with calibration pecks, which exact calibration settings we need to follow.
Consumer Products Guy wrote:
I'm retired now, and our 5890 units were hated by my last supervisor. But they worked great with the 7673 autosampler. Let's assume that the 9-pin remote cables are all in place...



juliomioto wrote:
Hi. Thank you for your response and your time.
So - the issue was the remote cable from GC was not connected to the controller. Now everything is working.


Ha !!! Old guys rule !!!!


juliomioto wrote:
We now need to work on the method for short-chain fatty acids.


What chain lengths? What matrix? What levels?
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry