- Posts: 4
- Joined: Sun Jan 12, 2014 6:07 pm
During feasibility study, in some injections (2 to 4 injections out of total 45 injections) we are getting a nice taller and narrow peaks with clean baseline (it's not an electrical noise, no retention time shift). Normally, alcohol peaks we see are twice shorter in height and wider with Helium carrier gas and N2 carrier gas. With He carrier gas we never saw taller and narrow peaks. To rule out the instrument, we tried 2 GC instrument and on both instrument we saw 2 to 4 injections with narrow, sharp peaks with N2 as carrier gas. We changed the liner, syringe and septa for every run.
In the method, we did not change any parameters (liner, temp, syringe etc.) when switched from He to N2 carrier gas. To rule out the possible causes, we don't have the Universal Trap for N2 gas, while for He we have a Universal Trap that's the only difference. We are not sure why this is happening only in some injections with N2 as carrier gas? Any ideas or suggestions to fix the issue?
GC Parameters:
Agilent GC 6980
FID Detector with capillary column
Split injector
Carrier and Make up gas: N2
Thank you so much!