- Posts: 53
- Joined: Mon Sep 26, 2011 4:02 pm
- Location: USA
New TCD detector baseline dip
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
My new TCD detector on 7890a is experiencing a uncharacteristic plunge at what would be the tail after my brominated compound peak but not after my phthalate peak. The plunge recovers pretty quickly. The plunge where the tail of the peak is almost looks like a smaller negative peak. The baseline recovers pretty soon. This only happened now after a new TCD was installed - this effect has seemed to mostly flatten out over the course of days and many injections of the material. It was messing up my quantitation due to the integration problem it causes trying to establish peak end of that brominated peak. I have never seen this before and need help figured this out!!
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
This one has my curiosity piqued.
You are running a bromo compound on a TCD ~ so relatively high levels since you can see it with a TCD.....
And you had it running successfully before? I would assume you are using helium carrier? And you are using a capillary column running up to how hot???
Best regards,
AICMM
You are running a bromo compound on a TCD ~ so relatively high levels since you can see it with a TCD.....
And you had it running successfully before? I would assume you are using helium carrier? And you are using a capillary column running up to how hot???
Best regards,
AICMM
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
Has this problem been resolved ? What column is being used?
Gasman
Gasman
-
- Posts: 53
- Joined: Mon Sep 26, 2011 4:02 pm
- Location: USA
AICMM wrote:
This one has my curiosity piqued.
You are running a bromo compound on a TCD ~ so relatively high levels since you can see it with a TCD.....
And you had it running successfully before? I would assume you are using helium carrier? And you are using a capillary column running up to how hot???
Best regards,
AICMM
Yes I have been running this method and conditions for years.
UPDATE:
The strange situation has resolved itself after working on the EPC controller connections. I think when replacing an EPC controller having it positioned exactly right is critical. It wasn't until I took the EPC out and reinstalled it that the problem resolved itself.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
The reason that I asked , is that some years ago we saw a similar problem with fatty acids, namely butyric acid. The Agilent TCD works on a switching principle where it switches at 5 Hz between the sample from the column and the carrier gas. The signal is the difference between the two gas streams. If sample "sticks" to the detector and is "slow" to leave the detector you will get a false signal. Agilent does supply a special modified detector to solve this problem.
Gasman
Gasman
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science