Hi there,
I am using my CP WAX 57CB column for a GC-FID with a 1.2 ml/min flow rate (He) and the pursue is very high 134 kPa. I removed the column and cut 1cm and connected the pressure was still high. The FID got to 2000 pA (did a blank injection).
I really appreciate if you help me to solve this issue.
Thanks in advance.
SherryGC-MS wrote:
Hi there,
I am using my CP WAX 57CB column for a GC-FID with a 1.2 ml/min flow rate (He) and the pursue is very high 134 kPa. I removed the column and cut 1cm and connected the pressure was still high. The FID got to 2000 pA (did a blank injection).
I really appreciate if you help me to solve this issue.
Thanks in advance.

Think of a column as a restrictor. Column ID and lenght is more important than type of a column.
Check if your column isn't clogged and whether it is properly installed.
What were the circumstances you noticed high pressure ?
Any maintinence works on inlet, column or detector.
What was the pressure earlier, before you noticed 134 kPa.
Is it configured properly in GC ?
Hey, that's first time I'm using this column so no information on previous pressure. I usually get 45 kPa on my TG-5 column (30mx 0.25x0.25) and thought 134 kPa is too high! However as I checked the pressure can go as high as 400kPa for a 50m column with 0.25 mm ID, but I still think 134kpa at 35C is too high!
I check the column connection and reconnected it. maybe my conditioning was not enough?
SherryGC-MS wrote:
Hey, that's first time I'm using this column so no information on previous pressure. I usually get 45 kPa on my TG-5 column (30mx 0.25x0.25) and thought 134 kPa is too high! However as I checked the pressure can go as high as 400kPa for a 50m column with 0.25 mm ID, but I still think 134kpa at 35C is too high!
I check the column connection and reconnected it. maybe my conditioning was not enough?

Hi. Conditioning influences baseline not pressure.
Check with: https://ez.restek.com/ezgc-mtfc what pressure should you get for your flows and temperatures. Isn't your column clogged with rests of rubber cap in the column ?
When trimming a column, take a chunk if you suspect that the end is clogged or dirty. The only reason to score and knock off 1CM is if you made a ragged cut when removing the previous 7-10CM.
134kpa sounds reasonable for a 60M column given your 30M column pressure.
Thanks,
DR
Image
If you double the column length you will more than double the pressure. Just a thought.
The past is there to guide us into the future, not to dwell in.
Thanks @dblux_ for sharing the Link. It is vey handy and makes life easier!
Thanks @DR I will cut 7cm.
Thanks @James, I have benefited from all your responses in GC related posts.

Looks like 110kPa is a normal pressure at 35C for a 50m column.
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry