I have a GC equipped with Mol Sieve 5A and a TCD and N2 as an carrier gas. I have tried different combination of oven and detector temperature but i am not able to detect CO. What conditions should i use to detect CO?
-
- Posts: 42
- Joined: Fri Jul 29, 2022 6:22 pm
Used to run ASTM D-1946/ EPA 3C. We used a GS-GasPro column, argon carrier gas and ran with cryo. Starting temp was -80.
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
szsayyed,
A mol-sieve should separate CO - typically later in the chromatogram - unless the column has been badly abused.
The bigger question here, I think, is at what level. Using nitrogen as a carrier, the sensitivity to CO will be poor - probably low percent realistically.
If you are looking that low or lower, you should consider switching to helium or hydrogen as the carrier in order to gain sensitivity.
Likewise, keep the detector cooler if you can - on the order of 175 or so.
Best regards,
AICMM
A mol-sieve should separate CO - typically later in the chromatogram - unless the column has been badly abused.
The bigger question here, I think, is at what level. Using nitrogen as a carrier, the sensitivity to CO will be poor - probably low percent realistically.
If you are looking that low or lower, you should consider switching to helium or hydrogen as the carrier in order to gain sensitivity.
Likewise, keep the detector cooler if you can - on the order of 175 or so.
Best regards,
AICMM
-
- Posts: 4
- Joined: Fri Jul 29, 2022 6:22 am
Dear sir,
I have tried injections with pure CO as well at TCD temperature 140 degrees. but even then theres no sign of a peak
Analysis is performed on the new column
I have tried injections with pure CO as well at TCD temperature 140 degrees. but even then theres no sign of a peak
Analysis is performed on the new column
-
- Posts: 348
- Joined: Mon Sep 13, 2010 8:28 am
szsayyed wrote:
Dear sir,
I have tried injections with pure CO as well at TCD temperature 140 degrees. but even then theres no sign of a peak
Analysis is performed on the new column
As was mentioned above - try lower temperature. Without cryo -80 is not reachable, start at +40 in oven.
Looks like your peak is very broad and not seen as a peak.
Also very hot TCD may decrease sensitivity, try +90.
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
szsayyed,
Let's start simpler. If you inject air do you get an early, good sized O2 peak?
Best regards,
AICMM
Let's start simpler. If you inject air do you get an early, good sized O2 peak?
Best regards,
AICMM
-
- Posts: 4
- Joined: Fri Jul 29, 2022 6:22 am
No sir i did not get an O2 peak at the beginning
-
- Posts: 4
- Joined: Fri Jul 29, 2022 6:22 am
AICMM wrote:
szsayyed,
Let's start simpler. If you inject air do you get an early, good sized O2 peak?
Best regards,
AICMM
I could observe a small peak at tha begining but not a huge one
-
- Posts: 1742
- Joined: Fri Feb 09, 2007 10:33 pm
- Location: Albq, NM
szayyed,
Which TCD are you using - what vendor. Some of the TCD's use range settings to moderate sensitivity so I am wondering if that might be an issue. The other think I would suggest, if possible, can you inject an aliquot of hydrogen and see what kind of response you get from that. That would certainly tell you if the TCD is dead since, against N2 carrier, it should give you a great peak.
Best regards,
AICMM
Which TCD are you using - what vendor. Some of the TCD's use range settings to moderate sensitivity so I am wondering if that might be an issue. The other think I would suggest, if possible, can you inject an aliquot of hydrogen and see what kind of response you get from that. That would certainly tell you if the TCD is dead since, against N2 carrier, it should give you a great peak.
Best regards,
AICMM
9 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science