How would the baseline signal for an ECD normally respond if you change the temperature?

I have two ECDs in one instrument. When I turn the detector heaters off, one signal dips a little bit and holds steady (a change from about 30 mV at 350C to about 25mV at room temp).

The signal from the other detector, though, actually increases. At 350C, it was showing 25 mV. At 120C, it was about 180 mV. If I hold the temperature there, the signal holds. If I turn the heater off, the signal gradually rises until it's at about 250 mV at room temperature.

Any thoughts?
MichaelVW wrote:
How would the baseline signal for an ECD normally respond if you change the temperature?

I have two ECDs in one instrument. When I turn the detector heaters off, one signal dips a little bit and holds steady (a change from about 30 mV at 350C to about 25mV at room temp).

The signal from the other detector, though, actually increases. At 350C, it was showing 25 mV. At 120C, it was about 180 mV. If I hold the temperature there, the signal holds. If I turn the heater off, the signal gradually rises until it's at about 250 mV at room temperature.

Any thoughts?


I could be a leak in the fitting for the detector. I have seen it before that the vespel/graphite ferrules, especially the larger ones that can be on makeup gas adaptor fittings on Agilents, will shrink over time with heating, and when the temperature drops they begin to leak. Usually snugging up the fitting nut solves the problem.
The past is there to guide us into the future, not to dwell in.
Thanks, that makes sense!

I was curious if Dust-Off would work as a leak detector on an ECD since it has a fluorinated compound, and tried giving the cooled-down detector side a squirt - sure enough, the signal spikes.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry