MSCHemist wrote:James_Ball wrote:MSCHemist wrote:
Yea that board is telling the GC there is 110 psi pressure in inlet A even when the inlet isn't hooked up to the board and at the same time telling it there is 119 psi in inlet B even though there is nothing and never has been anything hooked up to port B on that board. clearly it is shorted. I found a board on Ebay for < $100
except oddly today nothing is coming out the split/splitless vent and variable gas flow is coming out the septa purge vent.
If the board is bad, who knows what signals or commands it is sending to the actual inlet controls.
That is throwing me though. The septa purge regulator is just a mechanical 3 ml/min with no electronics. Not sure how I am getting higher flows out of that vent unless it is also governed by the PCV.
It is 3ml/min based on head pressure, since it is not adjustable, it will vary with head pressure. If the pressure is up, then the flow should be up some. I think also when the valve for split/splitless actuates some flow is diverted that way, but not certain since it has been quite some time since I looked at a flow diagram for those.