Hi all,
Our Scion 436-GC is having problems stabilizing column oven temperature, and the error 109 "column oven too slow response" is showed.
Anyone can help me with these problem?
Thanks a lot!
Sol
Sol
I don't know your instrument but when I've had this sort of thing happen on my Agilent systems, it usually means the flapper motor has failed. There's likely a flap of some sort in the back that keeps the hot air in or lets it out. It's not a difficult repair on a 5890/6890.
It could be an electronic, electrical or mechanical problem. Unless you have lots of time to spend troubleshooting it is probably worth asking a service engineer to call.

Peter
Peter Apps
Thanks for your answers. I think it is the column oven temperature probe that is damage. I'll change it and tell you if we solve the problem.
Sol
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry