What is the best flow for a column with the following dimensions?
30m*0.25mm*0.25μm

Thank you
I would start from 1.5 mL/min, but it could be between 1.2 and 1.8 mL/min, depending on your application
Thank you trozen
Optimum flow (giving the best resolution) depends on the carrier gas in use. With hydrogen & helium faster flows can be used with relatively little change in resolution, but column efficiency drops of fairly rapidly below the optimum.

As a rough guide a linear velocity of 10 cm/s is near optimum for nitrogen, helium's optimum is roughly twice this (20cm/s) & hydrogens near double again (40cm/s).

Look up 'Van Deemter' plots, for a fuller picture.