How to select "loop final pressure" on Agilent 7697A
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
How to select "loop final pressure" on Agilent 7697A headspace? Should the final loop pressure be higher or lower than the carrier pressure?
-
- Posts: 5
- Joined: Wed May 15, 2019 10:40 am
Final pressure is not related to carrier pressure. You should choose a pressure lower than that used to pressurize vial: this difference is used by system to fill loop with sample. initial pressure is dependent from sample matrix and method condition: for example you can start from 15 psi and finish at 11.
Try to find best combination for your samples.
Try to find best combination for your samples.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science