We are having issues passing our methaniser check. A standard with trace CO, CO2 and methane is run monthly, or after any maintenance event and the FID response (area counts) is recorded and trended. If the area is trending downwards or has changed by more than 5% change must be
investigated.
Basically we rarely passes, has anybody any experience in what we can do to solve this issue
Which component is more than 5% off spec.?
Or which component is trending downward?

Or do you mean this is happening for the total area of all peaks?

It could be that the catalyst in the methanizer is no longer converting all the components to methane. In that case the methanizer cartridge can be refilled or replaced.
seamoro wrote:
We are having issues passing our methaniser check. A standard with trace CO, CO2 and methane is run monthly, or after any maintenance event and the FID response (area counts) is recorded and trended. If the area is trending downwards or has changed by more than 5% change must be
investigated.
Basically we rarely passes, has anybody any experience in what we can do to solve this issue


Problem solved?
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry