Hello,

I have a new Agilent 8890 GC that has been giving me a lot of trouble considering it is a new instrument.

One thing I just noticed yesterday was that the instrument would not stabilize and kept saying make up gas not ready. I checked and my makeup was set to 2 mL/min but the actual flow was 2.4 mL/min. I can set the flow higher than the 2.4 and get the instrument to ready, or turn off the make up and get it to ready, but for whatever reason I cannot get the make up flow to go below the 2.4. Any ideas?
Does it specify a range that it will support? I don't think I have ever seen anything that uses only 2ml/min of makeup flow.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
Does it specify a range that it will support? I don't think I have ever seen anything that uses only 2ml/min of makeup flow.



The range is between 0 and 100. I agree the 2 seemed low but the method was developed by a custom solution group and the main thing I'm concerned with is that it was running at 2 ml/min just fine up until this week. Trying to determine what caused the change.

The unit is only about 6 months old and has been giving me a lot of trouble and this is just the latest example of unknown changes.
AFAIK Agilent flow regulators is actually a pressure controller with appropriate restrictor placed at output.
Probably it is not tightened well. Check your tubing.
jwf239,

A suggestion - turn off the hydrogen and air to the detector and see if the make-up can get to it's set point.

Best regards,

AICMM
JWF239 wrote:
James_Ball wrote:
Does it specify a range that it will support? I don't think I have ever seen anything that uses only 2ml/min of makeup flow.



The range is between 0 and 100. I agree the 2 seemed low but the method was developed by a custom solution group and the main thing I'm concerned with is that it was running at 2 ml/min just fine up until this week. Trying to determine what caused the change.

The unit is only about 6 months old and has been giving me a lot of trouble and this is just the latest example of unknown changes.


You are stressing over a pointless quirk in a method. Presuming that this is an instrument with an FID, 2 ml/min of makeup flow will make no difference at all to anything, as long as the column is properly installed. How long was the warranty on the instrument ?
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry