Problem with TD-20 thermal desorption unit
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
We have a problem with TD-20 thermal desorption unit (Shimadzu). When we try to run samples, it appeared a message 'Desorber error (Carrier) occurred'. At the same time, sampling Tenax tube was pressed too tight to support. As a result, the positioning engine moved constantly and made a strange sound. Did someone have the same problem, and what were the possibilities to solve it?
-
- Posts: 24
- Joined: Fri Apr 30, 2021 1:10 pm
- Location: Italy
Change desorber. It has a lot of problems. I had it for years.
The most important problems is the absence of a electronic leak check at each run, so that if you have a leak you cannot know it and you fail the analysis.
The most important problems is the absence of a electronic leak check at each run, so that if you have a leak you cannot know it and you fail the analysis.
-
- Posts: 3
- Joined: Fri Jun 11, 2021 5:01 pm
CrisTheChemist wrote:
Change desorber. It has a lot of problems. I had it for years.
The most important problems is the absence of a electronic leak check at each run, so that if you have a leak you cannot know it and you fail the analysis.
Thanks for the answer. It is basically what we did. It was replaced with TD-30R.
-
- Posts: 24
- Joined: Fri Apr 30, 2021 1:10 pm
- Location: Italy
Very good.
Can I ask you how does it works? Do you are satisfied of the TD-30? Are there some troubles?
I have a Perkin Elmer desorber now.
thank you
Can I ask you how does it works? Do you are satisfied of the TD-30? Are there some troubles?
I have a Perkin Elmer desorber now.
thank you
-
- Posts: 3
- Joined: Fri Jun 11, 2021 5:01 pm
CrisTheChemist wrote:
Very good.
Can I ask you how does it works? Do you are satisfied of the TD-30? Are there some troubles?
I have a Perkin Elmer desorber now.
thank you
We got TD-30R. It's better than TD-20. Two main advantages are possibility to restore samples and leak check. It's all easier to change O-rings. Troubles... Some issues with TD and GC-MS software interaction. It damages O-rings quicker than TD-20.
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science