-
- Posts: 17
- Joined: Sun Mar 17, 2019 6:57 am
equivalent column for COS in Propylene
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
Agilent says that to avoid co-elution of COS and Propylene , "select low sulfur column" (CP8575) should be used. I wanted to know what the equivalent column from RESTEK or SCION is. ASTM D6228 complies with this analysis
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
Mreza555 wrote:
Agilent says that to avoid co-elution of COS and Propylene , "select low sulfur column" (CP8575) should be used. I wanted to know what the equivalent column from RESTEK or SCION is. ASTM D6228 complies with this analysis
If you read paragraph 6.2.3 of the method, you will see that other columns can be used.
6.2.3 The column shall provide adequate retention and
resolution characteristics under the experimental conditions
described in 7.3. One example of a capillary column and
operating conditions used with the FPD is shown in Table 1.
Two examples of columns and operating conditions used with
the PFPD are shown in Table 2. Other columns, which can
provide equivalent separation, can be used as well.
Gasman
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
Mreza555 wrote:
Agilent says that to avoid co-elution of COS and Propylene , "select low sulfur column" (CP8575) should be used. I wanted to know what the equivalent column from RESTEK or SCION is. ASTM D6228 complies with this analysis
If you read paragraph 6.2.3 of the method, you will see that other columns can be used.
6.2.3 The column shall provide adequate retention and
resolution characteristics under the experimental conditions
described in 7.3. One example of a capillary column and
operating conditions used with the FPD is shown in Table 1.
Two examples of columns and operating conditions used with
the PFPD are shown in Table 2. Other columns, which can
provide equivalent separation, can be used as well.
Gasman
4 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science