dblux_ » Sat Mar 06, 2021 11:53 am
Fouadm wrote:
Hi
Many thanks to you in advance.
I have a problem with the front FID which do not go on as the igniting trials have not succeeded.
I have a doubt about the front inlet configuration.
Any suggestions are highly appreciated.
1) Does your igniter glow ?
2) Check if hydrogen and air go to FID. Don't rely on setting values. You must prove by flowmeter or by other means. This will check if FID nozzle is not clogged as well.
3) Set stoichiometric flows of gases. Reduce makeup and column flow.
4) Blow a little when igniting.
It must ignite.
BTW - let hydrogen flow as long as air is blown from the line.