Asking for help/suggestions on figuring out why chemstation does not start data acquisition after injector injects.... Here are the details.....
5890 Series II
Dual FID
7673 Injector
HPIB cabling
version B.04.03
Basically all is fine in the operation of GC and autosampler, but the last step, the most important one, when the autosampler injects, instead of Chemstation starting the run (data acquisition), immediately after the injection it goes to POST RUN...... In other words, it seems to skip the data acquisition step. The GC does not start either.
SIDE NOTE: If I push START manually in the GC, then GC starts the run and data acquisition also starts data acquisition after being in POST RUN mode
I have tried different autosampler towers, different controller boxes, different HPIB cables and also another computer with a different version of Chemstation. Get the same problem.
Please advise on any other modifications I can make or maybe you have had this same problem before. Thanks in advance for any comments on this.