AZBiodiesel wrote:
Hi what could cause N2 makeup gas on an agilent dual FID 6890 GC not reach setpoint? Setpoint is 20mL/min for rear detector but its showing 0.2 regardless of pressure and set point increase. Front detector makeup gas hits setpoint no problem. GC is getting proper pressure (70psi) confirmed good flow to the rear inlet ports.
I just removed FID assembly,removed jet and inspected, replaced the large FID graphite seal, re cut and inserted column, removed 1mm from Jet tip and tightened column nut ferrule. Still no flow. Is there somewhere it could be plugged?
Detector lights no problem and I am getting good chromatography, just curious. This detector is used with an on column injector if that makes a difference?
Thanks,
Can you measure any flow at the outlet with a flow meter?
Since it is reading only 0.2 on the instrument panel, then it could be that the EPC is bad. We have had to have those replaced in the past. There are a few companies that will rebuild them for a lot less than Agilent wants for new ones.
If the instrument panel was reading flow but nothing at the detector then I would suspect a broken line, usually at the makeup gas fitting. Since they measure the flow at the flow controller then a low reading is either bad flow sensor or no flow getting through the control solenoids.