hi everyone
i have a gas CRM with these description:

component name: hydrogen
concentration (moles) :4.9 %
accuracy(+/- %):2

now i want to calibrate my GC with this CRM, i calculate the uncertainty in this way:

u=d/√3 u=2/1.73= 1.154

so when i inject the CRM to GC, my results could be in 4.9 ± 1.154 range

is it right?
Hi,

The range calculated by your formula is too high (>20%) so I don't think the formula is correct. In fact if you have the access to GPA2261 the uncertainty is listed inside and has been widely accepted as the square root of (CRM uncertainty ^ 2 + instrument reproducibility ^ 2). This formula works much better as it takes into account the instrument drift which definitely affects the CRM reading when running on your GC. Reproducibility often can be found in a standard method if you are following any, otherwise you could easily google how to estimate it.

Hope this helps.

Peter.
thanks for your answer
i found this method in this study:

A Practical Guide to the Calculation of Uncertainty of Measurement
Mario Zilli*
ARPA F.V.G., Settore Laboratorio Unico, Laboratorio di Gorizia, Italy


https://imgur.com/ZX8DRnR
peter_tran wrote:
Hi,

The range calculated by your formula is too high (>20%) so I don't think the formula is correct. In fact if you have the access to GPA2261 the uncertainty is listed inside and has been widely accepted as the square root of (CRM uncertainty ^ 2 + instrument reproducibility ^ 2). This formula works much better as it takes into account the instrument drift which definitely affects the CRM reading when running on your GC. Reproducibility often can be found in a standard method if you are following any, otherwise you could easily google how to estimate it.

Hope this helps.

Peter.

thanks for your answer
i found this method in this study:

A Practical Guide to the Calculation of Uncertainty of Measurement
Mario Zilli*
ARPA F.V.G., Settore Laboratorio Unico, Laboratorio di Gorizia, Italy


https://imgur.com/ZX8DRnR
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry