by
Laura._ » Fri Aug 21, 2020 11:50 am
Also when I start run without injecting solution, it appeared.
I assume that it shouldnt be caused by column, because it was changed by service technician and its new ( same peak was eluted using old column).
I use "old method", I didnt have problem with it until now.... so it shouldnt be caused by insuffiecient hold time, I think...
James_Ball wrote:
Does this peak appear if you start the run but inject nothing? You can remove the syringe and do a run that way so it doesn't inject anything. If it still appears then it is some residue that is adsorbing onto the filament and evaporating off once the filament heats up.
Cleaning the source will only help if there are some carbon deposits near the filament adsorbing the compound. How old is the column? Could it be the stationary phase is beginning to break down and bleed off onto the source?
Another cause could be a high boiling compound that is being trapped on the column at the end of the run and coming off at the beginning of the next run. If that is happening then you need to extend the hold time at the end of the analytical run to elute that compound before the oven cools down again.