I have read that Hydrogen is safe to use in Headspace autosamplers. But a co-worker brought up to me an Agilent note/document that indicates not to use Hydrogen on the Headspace A/S. So we are butting heads. It seems like co-worker is willing to settle for Nitrogen but I have not seen literature about Nitrogen being used in headspace. If any of you have any experience or knowledge on the subject please comment. We are trying to meet half-way.
-
- Posts: 139
- Joined: Sat Feb 27, 2010 5:27 am
According to Agilent, the gas used for the headspace sampler should be the same as the GC carrier gas. The Agilent 7697A can use hydrogen with a recommended purity of 99.9995% purity. I guess it just depends on what headspace sampler you have. When in doubt, nitrogen is also very cheap and does not pose a fire hazard. For me personally, I would just go with nitrogen and use brass gas lines.
https://www.agilent.com/cs/library/site ... _Guide.pdf
It also states to never use flammable gases for vial pressurization. So just go with Nitrogen.
https://www.agilent.com/cs/library/site ... _Guide.pdf
It also states to never use flammable gases for vial pressurization. So just go with Nitrogen.
-
- Posts: 32
- Joined: Mon Jul 03, 2017 6:25 pm
Pacerlaser,
thx for your comments. I think we are going to try Nitrogen.
thx for your comments. I think we are going to try Nitrogen.
-
- Posts: 391
- Joined: Tue Dec 30, 2008 2:57 pm
Yeah the danger is related to vial pressurization, you don't want the hydrogen going into the vial oven. Nitrogen is a better choice.
-
- Posts: 12
- Joined: Wed Oct 23, 2019 2:31 pm
- Location: UK
I've used both Helium and Nitrogen as carrier for Headspace but would never use Hydrogen.
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science