I guess I'm not clear on what you're saying. Is your end goal polymer identification? It might be helpful if you could explain in more detail, what you're trying to accomplish in your measurement.
When you use the term "fragment" in the context of pyrolysis sampling, it could mean how the polymer breaks apart upon intense thermal degradation. In mass spectrometric terms, it means those mass fragments that result because of the sampling introduction into the source of the detector.
If you are analyzing BPA-based polymers, thermal degradation should result in a lot of isopropylphenol and phenol (relatively much smaller amounts of methylphenol). If it is a BPF-based polymer, thermal degradation will result in a lot of phenol and methylphenol (very little isopropylphenol). I believe that Acetone is used to make the bridging carbon between the phenol rings in bisphenolA. That's why it's an "A". Formaldehyde is used to make the bridging carbon in bisphenolF. That's why it's an "F". It is not inconceivable that you might have some residual formaldehyde or acetone in either process and end up with a little bit of one or the other in every sample.
I doubt there's a purely BPA or BPF polymer that you can find, anywhere.
Under pyrolysis conditions, you should be able to use the relative amounts of isopropylphenol, phenol, and methylphenol to sort out if the polymer is F-based or A-based. One or the other pair will predominate in your pyrolysis GCMS traces.
It would be a good trick to make bisphenol-A from bisphenol-F. BPF is a simpler molecule. Is there a way to make BPF from BPA? Perhaps but it seems highly unlikely. Knowing the structures of these epoxy polymers, I just can't see a good synthetic reason for why the methyl groups would leave preferentially and be substituted by hydrogen atoms.
The mass spectra of BPA and BPF are clearly different from each other. Even if they coelute, you should be able to sort your TIC by 107, 183, and 200 for BPF and 119, 223, 228 for BPA and tell the difference. 228-213 = 15 (loss of a methyl group). Phenols typically give great molecular ions in the mass spectrometric detector.