I am trying to work out the total carrier gas flow used in my GC methods. I have a 7890 with two inlets, each method using one inlet only while the other inlet is “idle”. Front inlet is connected to a FID, Back one to a MSD. Each inlet is configured in split or splitless depending on the method.
I am trying to get my head around the purge to split flow, the gas saver flow..and am confused! So just looking for someone to double check that I have this right. Here is an example of the total gas flow calculated for a method where the front inlet is used for sampling in Split mode whereas the back inlet is not used:
Inlet Front
Injection Split
Split ratio 20:1
Column flow ml/min 1.5
Column+Split flow ml/min 31.5
Septum purge ml/min 3
Total ml/min for inlet 34.5
Inlet Back
Injection Splitless
Column flow ml/min 1
Septum purge ml/min 3
Purge to split ml/min (1 min) 20
Total ml/min for inlet 24
Total ml/min for both inlets 58.5
After 2 minutes, the gas saver leads to:
Inlet Front
Injection Split
Column flow ml/min 1.5
Gas saver ml/min 20
Septum purge ml/min 3
Total ml/min for inlet 24.5
Inlet Back
Injection Splitless
Column flow ml/min 1
Gas saver ml/min 20
Septum purge ml/min 3
Total ml/min for inlet 24
Total ml/min for both inlets 48.5
Does this look ok?
Also, what is best for an inlet that is not used (back one in my example)? Do I really need a septum purge and a purge to split vent?
Thanks for the help