Unexpected SPME-GC/MS chromatogram

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi all

We are doing GC-MS analysis of manually SPME-extracted volatiles from fermented milk. There are three issues with this:
1. The peaks of our interest as analyzed by MS have weird shapes as indicated by arrows in Fig. A
2. There is a huge peak at the beginning of all the chromatograms including that with blank fiber as indicated in Fig B. The mass spectra in this area match with those of water, nitrogen, etc. This peak appears only with manual SPME and not with autosampler injection of liquid standards.
3. There are many peaks of the matrix (siloxane) in the SPME chromatograms and not standards’ chromatograms (liquid injections), meaning there is bleeding from SPME matrix. We do condition the SPME fiber for 30 min in the injector before headspace extraction. The fiber has been used only for about 10-20 times till now.

Can anyone recommend any remedies to troubleshoot these issues?

Many thanks

Image
https://postimg.cc/vx1YY8N8
Your peaks, mainly the early ones are not focusing well.
You can reduce the desorption time
you can reduce the GC start temperature as low as 35 deg C
You can use a thicker phase column

Sometimes peaks focus better on a polar wax coln or a nonpolar db-5 column

The air void peak is normal
siloxane peaks are also normal not much you can do but ignore them
Also for samples with a lot of water, you could add something like a Restek Hydroguard guard column at the head of the column to help focus the water and give it a sharper peak.

Are you interested in water and nitrogen or any analyte that has a mass below m/z35, if not start the scan range higher and those will be ignored.
The past is there to guide us into the future, not to dwell in.
Thank you MSCHemist and James_Ball for your replies. We were taking the earlier run under splitless mode. Changing to 1:20 split has solved the problem to a large extent.

Thanks for your time!
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry