Agilent 7697A headspace carrier control by HS?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I am using an Agilent 7890 GC with 7697A headspace unit. Currently we are using GC carrier control, but we have a G4562A carrier gas PCM module just sitting in a drawer. What advantages would carrier control by HS or GC+HS provide?
It looks like Agilent recommends carrier control by GC only when using the 7890, so I guess we will stay with our current configuration.
You do not mention if you are using a capillary inlet. If you are, then you should use the GC to control the gas flow, then the GC knows what flow to set up for the split ratio that you require. If the flow source is external to the GC, you will not know the split ratio.

Gasman
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry