I am using an Agilent 7890 GC with 7697A headspace unit. Currently we are using GC carrier control, but we have a G4562A carrier gas PCM module just sitting in a drawer. What advantages would carrier control by HS or GC+HS provide?
-
- Posts: 245
- Joined: Sat Apr 28, 2007 3:24 pm
It looks like Agilent recommends carrier control by GC only when using the 7890, so I guess we will stay with our current configuration.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
You do not mention if you are using a capillary inlet. If you are, then you should use the GC to control the gas flow, then the GC knows what flow to set up for the split ratio that you require. If the flow source is external to the GC, you will not know the split ratio.
Gasman
Gasman
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science