GC-TCD issues

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Good evening all,

I am hoping to get some advice on how to go about optimising analysis using GC-TCD (Varian CP3800), Helium carrier gas.

I have a method set up for analysis of methane, ethane and CO2. I have a valve switch at 1.6 min that retains Oxygen, Nitrogen and Hydrogen on my molsieve column. Methane is eluted from my Haysep Q column at 1.7 min, CO2 at 2.8 min and ethane 5.8 min. A valve switch at 8.3 min is followed by Oxygen and Nitrogen eluting from the molsieve column at 9.5 and 10 min respectively.

I will need to tweak the protocol to allow for the analysis of propane (when standards arrive). My main issue is having spent a week setting up this protocol, my supervisor is hoping that my limit of detection for ethane is around 500 ppm. Currently, I cannot detect ethane at levels lower than 1000 ppm. As this is my first experience of using GC-TCD I am not really sure if there is a way to improve my limit of detection with my existing columns. If someone could confirm that this is the case, or suggest a different option I would be very grateful.

My second issue is, having set up the above protocol I have been asked if I will be able to calibrate for H2. My understanding is that this will be impossible using Helium as a carrier gas.

I have access to Argon as a carrier gas, which would make H2 analysis possible, however am I naive in thinking I could just switch carrier gases from one to to the next depending on whether I need to analyse H2 or alkane gases?

With thanks,
Rich


Image
https://imgur.com/a/oO9o3Vk
You can change sample loop - if you have 1cm3 loop for example, try to replace this loop by 2cm3 loop. Or you can increase bridge current. And optimal gas flow.
Apologies for the late reply. I hope you were able to figure out the method development issues. My company has years of experience with GC-TCDs and hydrocarbon/natural gas analyses. If you ever encounter future problems, keep us in mind and would gladly provide a consultation.

service@aasystems.com
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry