I work with GC system 7890B ECD we recently installed the Mass hunter software version 10.0 368 and I'm curious to know if any one is experiencing issues with calculations for data reporting. I've ran several QC's from outside vendors and I'm experiencing issues where my undiluted samples report lower than my Diluted samples. I work with Haloacetic acids.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
Are you putting a dilution factor in the sequence table for the diluted samples?
When I do it for GCMS samples what shows in the Quantitative Analysis batch screen is the on column value but in the report file for uploading to our LIMS it has the on column value multiplied by the dilution factor.
I think this is so the value is compared directly to the upper and lower calibration level for the blue or red flag on the concentration if it exceeds those limits.
When I do it for GCMS samples what shows in the Quantitative Analysis batch screen is the on column value but in the report file for uploading to our LIMS it has the on column value multiplied by the dilution factor.
I think this is so the value is compared directly to the upper and lower calibration level for the blue or red flag on the concentration if it exceeds those limits.
The past is there to guide us into the future, not to dwell in.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science