- Posts: 68
- Joined: Sat Feb 12, 2011 1:31 pm
Advertisement
NPD Agilent 7890/6890
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
Can I install NPD of Agilent 6890 onto Agilent 7890A?
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
No. The detector electronics board and the electronic flow controller for the detector are both different for the 6890 and the 7890.
Gasman
Gasman
-
- Posts: 1774
- Joined: Fri Feb 09, 2007 10:33 pm
push,
A minor point, but I suspect you can transfer the detector body itself and the pallet over. Then you need the proper 7890 EPC, electrometer and input board so the cost savings are not going to be very extensive at all..... I would advise buying the whole kit from Agilent if you want to do this.
Best regards,
AICMM
A minor point, but I suspect you can transfer the detector body itself and the pallet over. Then you need the proper 7890 EPC, electrometer and input board so the cost savings are not going to be very extensive at all..... I would advise buying the whole kit from Agilent if you want to do this.
Best regards,
AICMM
3 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Bing [Bot] and 10 guests
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Bing [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science