The problem with the compound decomposition

Hi,
I hope somebody could give me a help. Now I am using GC-FID to test the one compound purity. But I found the problem:
The compound boiling point is 102 C. it is very easy to decompose to other two compounds.
recently the method was set as following:
Initial rate temp wait
50 3.0
15 150 0
30 220 0
Inject port temp 105C. detect port temp 150C
Split Ratio is 46:1,
But at this condition, still small parts of compound had been decomposed.
Irene-

Perhaps GC is not the best methodology here. Is your compound better suited for HPLC? You did not post enough information (sample prep/matrix, sensitivity required, etc.) for more thorough help in your problem.

An additional thought, if you know the identity of your decompostion products and can obtain standards of them, maybe you could quantitate your peak of interest via mass balance.

Just my two cents.

-Jeff

You may try cool on-column injection, i.e. injection at a relatively low temperature. You may modify the split/splitless injector to on-column injection by an appropriate tapered liner.
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry