Advertisement

Alternative Column for Glass Column Packed

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
I have a column with the following specifications
Zebron ZB-50, GC cap, column 30 m x 0.32 mm x 0.50 um (G17)
I want to analyze using GC based on the compendial method, in the reference the column used is 1 m x 4 mm, glass column packed with 5% liquid phase G17 on 100 - 120 mesh support S1AB. Are the two columns equivalent? If not, is there a column recommendation that is equivalent to the column used in the reference?

Thank you for your time.
I would've thought glass columns were obsolete, I hadn't used those since about 1980 when fused-silica capillaries became available.

Of course, compendial methods seem to rarely change...
I also think that it is very rare to use the glass packed column method. But in terms of the regulations in our place, we have to be able to explain the reason for not using the column. I need references that can be used as justification for using the capillary column as an alternative.
Who enforces the regulations? They must have a lab that runs the test. Can you just call those people and ask?
They must have a lab that runs the test.
Ha !!! I wouldn't bet on that.

Would the two columns perform the same, as both are G17? No, the capillary would likely be better.


The original poster did not exactly state that he was assaying something that needed to be for a regulated product (e.g. cGMP); but if so, the QA department needs to get some documentation on file stating their rationale for such change. USP <621> even lists some changes which do not need validation (unless one has a pointy-haired boss like I had, who didn't believe what the 26 letters formed on the page).
The product will be marketed to SEA region and USP regulation is used. I asked the USP and they advised me to do a full validation. Thanks for the help.
6 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry