HI everyone,

I am currently trying to set up method 522 for analysis of 1,4-dioxane in SIM mode. The main issue I am having is high background for ion 88, the QI for 1,4-dioxane. How can I decrease the baseline of the 88 ion to get better detection at lower concentrations of my analyte?
QI is a pretty garbage abbreviation when quant and qual are both Q's. I've seen coelutions on 58 but 88 is usually just a falling baseline. What column are you running?
My mistake, 88 is the quantitative ion.
The column used is a 30mx0.25mmx1.4um thickness RXI-624Sil MS
Are you seeing the same baseline issue in your blanks? Have you tried re-shooting on scan and seeing if anything pops up as a TIC? We're running a CP Select 624 CB with the same dimensions down to a 0.05 ug/L reporting limit.
Yes even when running blanks in SIM mode, the 88 baseline is still high.Full scan blank runs have also shown the same falling baseline for the 88 ion. We hypothesized that it could just be a result of DCM causing the high background but even when keeping the oven at 30 C for 5 minutes (where DCM elutes out fully in blank DCM run) the high background is still present.
Not sure if you're going to be able to escape the falling baseline. Our integrator consistently takes an extra chunk of baseline on low points but since its baked into the calibration the quant values are fine. Tried setting it up to go valley to valley for awhile but it would cut the IS wonky on high concentration samples so we ditched it.

Image
70 eV wrote:
Not sure if you're going to be able to escape the falling baseline. Our integrator consistently takes an extra chunk of baseline on low points but since its baked into the calibration the quant values are fine. Tried setting it up to go valley to valley for awhile but it would cut the IS wonky on high concentration samples so we ditched it.

Image


What is the set up for you SIM? Dwell times, starts of segments, and gain factor?
We run a half baked version of 522 since it never came under federal drinking water regulations. It's just 1,4-D and d-1,4-D as the IS. 4 ions total, single group, 100 ms each. SIM gain of 2, ~1450 volts and the multiplier is at least three years old. Tunes around 1250 and only cleaning it sees is the yearly PM. Runs ~800 samples a year so most of the time it's sittin around big chillin.
Image

This is currently what 2ppb standard looks like with an EM Voltage at 1525. We need to be able to detect 0.3 ug/L. What can I adjust to get this peak to be more prominent/ decrease the baseline noise.

https://www.dropbox.com/scl/fi/uxmvb8pv ... vq8rx&dl=0
Our low point is 5 ppb, are you accounting for the 100x concentration factor that a sample would have? That would put you at 0.02 ppb.
Appears using a 2mm liner seem to have fixed the signal issues being low instead of using 4mm liners. Signal looks to have improved significantly with this change.
Are you running splitless or a split mode injection?

Sometimes you can run split 1:5 and have almost the same sensitivity as splitless and lose a lot of the tailing of the solvent. Also if splitless, try shortening the splitless hold time until you see a decrease in the target peak, holding it too long will transfer more of the solvent to the column causing the tailing off that you see.
The past is there to guide us into the future, not to dwell in.
70 eV wrote:
Our low point is 5 ppb, are you accounting for the 100x concentration factor that a sample would have? That would put you at 0.02 ppb.



I believe the method (Option 2) is 100 mL sample, concentrated to 2 mL of eluent, so only a 50x concentration factor. Right?

We see a similar falling baseline.

Are your high baseline/backgrounds only observed in the extracted samples, or both the calibrators and extracted samples?
13 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry