Hello everyone,

I have to build a method for the quantification of benzene in plastics.
I have agilent GC-MS-HS, but i dont know how to organize the standard to build a calibration curve.
I know someone use a mix of benzene in triacetin and make 4 dilution for the calibration.

Any other experiences?
We have our domestic method standardized for air quality control in working places.

Calibration mixture is an air sample made in:
10 L glass bottle with 5-10 aluminium foil balls.
Every analyte introduced in known amount (1-5 mkl) with chromatographic syringe.
Closed bottle is shaked for 10 min at 20C.

Probe is injected with gas syringe.
You're seemingly tied to the static-headspace method so you have a few things to determine before you'll have a good method.

Triacetin is probably a good choice of a solvent because it's chemically similar to your sample.

If I have your problem and your equipment (static headspace sampling is an equilibrium technique), I would use the method of standard addition to calibrate. Make a benzene-in-triacetin standard, then make known additions to the same mass of PET (in different vials of course). The key is you'll have to "cook the benzene into the sample" after you add it to the vial to ensure you're at equilibrium. To do this, I make several vials with a known and the same amount of PET in them. I then add a known amount (the same) of the standard to each vial and seal them. Then I heat the vials to a temperature above the glass transition temperature of the solid sample for many different times. Analyze an unequilibrated vial as a the t=0 sample. Then analyze the samples at 0.5, 1, 1.5, 2 hours etc. and see how the benzene peak area decreases with time. Once it doesn't decrease anymore, you're at equilibrium. As the sample approaches equilibrium, the amount of benzene that's actually in the gas phase will be less than what was present at t=0.

At this point, you know what it takes to get your fortified samples to equilibrium. Now, you an just perform the method-of-standard addition to determine benzene in your samples. If you can verify that the calibration data is independent of the matrix, then you can probably buy some clean PET and use that as your universal calibration medium.

It may take less time to get your samples (no fortification with benzene) to equilibrium. You'll need to determine this too.

There's a great book out there by Ettre and Kolb called "Static Headspace Gas Chromatography" that discusses all of these techniques. I'd get a copy of that if you don't have it. The PerkinElmer app note employs the technique called Multiple Headspace Extraction (MHE) that reduces/eliminates the matrix dependence of the analysis. You may explore that as well. Perhaps that's why they heated the sample to 200 °C during their analysis.

https://resources.perkinelmer.com/lab-s ... 276907.pdf
Thanks everyone,

The i've already an instrumental method i heat samples at 200°C but i dont know how to do the calibration curve.

I should purchase benzene standard and make 4 dilution starting from the mother one, so 4 point of calibration. But exist a std benzene in triacetine?
Well that's a bit of an open-ended question. If you make your additions and you do as I describe to make sure the data you collect is accurate, then you simply plot concentration of benzene on the x-axis and instrument response on the y-axis. Standard regression statistics apply. You can use a spreadsheet package like MS-Excel to do this.

If you're asking how do I do it in my software so I get a nice neat report printed out after my GC run, I can't tell you that because I don't know your software. All of the software manufacturers have their ways of doing things. If that's your question, you might check youtube.com. Lots of times, people post tutorials about this sort of thing.
5 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry