I am having an issue trying to draw up a 0.2 uL sample with a salvaged CTC A200 autosampler. I can easily draw that amount consistently with a hydrocarbon matrix but when I try to draw that with water I usually get nothing.
I switched to a gas tight syringe but that did not help.

I am using 50/50 xylene acetone for the rinse solvent. I have tried pre-rinsing the syringe with water sample as well as only pumping sample (without pre-rinsing) but that does not seem to do it either.

If I manually sample, I can pull a water sample, just not on the CTC.

Suggestions?

AICMM
* Decrease drawing speed
* Change syringe (needle tip may be clogged)
* Use "sandwich" - i.e. air gaps before and after sample
Would an intermediate rinse with Methanol or Acetone be possible to remove any traces of the xylene that may be causing a problem with the water entering the syringe?
The past is there to guide us into the future, not to dwell in.
Probably the vacumm created by a 0.2 ul movement of the pluger is not enough to break the surface tension of the water across the bore of the needle. Can you overfill and then eject all but 0.2 ul with that sampler ?
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry