Swagelok carrier gas valves- lubricated or non-lubricated?
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
Hello, I'm planning on installing 3-way ball valves just upstream of my GC-MS carrier gas inlets to allow for toggling between helium and nitrogen. Swagelok valves typically come pre-lubricated, though they do make a special non-lubricated version of all of their ball valves which is more expensive and has lower pressure ratings. I don't know if the silicone-based lubricant on their "normal" valve assemblies is something to worry about or not. Agilent technicians I spoke to did not seem to be aware there are versions with and without lubrication and defaulted to recommending the non-lubricated version because obviously, lubricants could be a source of contamination. Does anyone have any real-world experience with this?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
I have been using them to switch between helium and hydrogen on my 7000C MSMS and have not see any noticeable contamination. I used the ones sold by Restek https://www.restek.com/en/products/acce ... ves/46983/
These are the 3-way valves, so one out to the EPC and one in each from Helium and Hydrogen. Not sure if they are the lubricated ones or not. If there is any problem with contamination you could put a hydrocarbon trap after the valve and before the EPC just in case.
These are the 3-way valves, so one out to the EPC and one in each from Helium and Hydrogen. Not sure if they are the lubricated ones or not. If there is any problem with contamination you could put a hydrocarbon trap after the valve and before the EPC just in case.
The past is there to guide us into the future, not to dwell in.
2 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 16 guests
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm
Users browsing this forum: Semrush [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science