Hello! My first time posting a topic.

I am developing a method on headspace (Shimadzu AOC 6000 Plus) gas chromatography (Shimadzu Nexis 2030) with FID detector. I have to develope, optimize and validate a method that will analyse 16 different analytes (polar and non polar) in sludge.

RSD/% of Areas for Metoxy-2-propanol, Etoxy-2-propanol, Ethyl-benzene, m-,p-Xylene, o-Xylene are big (later eluting analytes with high boiling point).

For example:

I made a solution which consists of:

Analyte ______________V/V volume / volume (%)
Acetone_______________ 3%
Ethyl-acetate___________ 8%
Metoxy-2-propanol______ 10%
Etoxy-2-propanol________ 20%
Ethyl-benzene__________ 20%
Xylene________________ 39%

10 vials ( vial volume =20 mL) with 800 uL of the prepared solution are analyzed.

Conditions
Incubation temp: 80°C
Incubation time: 30 min
Syringe temp:150°C
Fill strokes count: 0
Injection flow rate: 3mL/min
Injection volume: 250 ul
Inlet temp: 200°C
Carrier gas: Nitrogen
Column: SH-I-624Sil MS 60 m x 0,25 mmID x 1,40 um df
Total flow: 154,5 mL/min
Purge flow: 3 mL/min
Column flow: 1,5 mL/min
Linear velocity: 27,6 cm/s
Split ratio: 1:100
Oven temperature program:
----------- 35°C 17,5 min
10°C/min 100°C 3,5 min
40°C/min 240°C 1 min
Total program time: 32,00 min
Detector temp: 240°C

Analyte ______________ RSD (%)
Acetone_______________ 5,80%
Ethyl-acetate___________ 10,32%
Metoxy-2-propanol______ 25,39%
Etoxy-2-propanol________ 23,68%
Ethyl-benzene__________ 23,19%
m-,p-Xylene____________ 23,24%
o-Xylene_______________ 23,65%

Why is RSD so high for later eluting analytes (with high boiling point)? I already excluded a possibilty of leaky vials (used new caps with new septums).

Any suggestions?
Try increasindg the volume of the standard - fomr 0.8 ml to 5 ml say.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry