- Posts: 138
- Joined: Sat Feb 27, 2010 5:27 am
Cannabis Residual Solvents math question
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
I am trying to set up a headspace evaporation residual solvent method using a GC/MS but I an trying to figure out how to calculate the final concentration of the cannabis residual solvent. For starters, I am using 200 mg (0.200 g) of plant material and adding 1 mL of 2,2-dimethylacetamide (DMA). From that mixture, I take 5 uL and add it to a 20 mL headspace vial and analyze it. Is the final calculation the instrument value in ug/mL multiplied by 0.005 mL (5 uL) divided by the 0.200 g weight if I want the units in ug/g?
-
- Posts: 159
- Joined: Thu Jan 05, 2012 5:27 am
pacerlaser wrote:
For starters, I am using 200 mg (0.200 g) of plant material and adding 1 mL of 2,2-dimethylacetamide (DMA). From that mixture, I take 5 uL and add it to a 20 mL headspace vial and analyze it.
Are you dissolving your analyte? It's not flower, because there's no reason to look for solvents in vegetable material. So you must be looking at concentrates.
1 mL of DMA + 200 mg of dissolved concentrates is going to give you a different volume, increasing your error substantially.
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science