I have issue with HP 7890A SCD355 for sulfur detection ,the issue is instability in result in sulfer mixture with helium balance , the result decreases and show big deviation along with runs

One more issue ,is there any formula to calculate sulfur conc. Of diffrent sulfur compunds by calibration only of one compound ( for example calculate conc. Of methyl and ethyl mercaptan by using gas STD. With hydrogen sulfide with nitrogen balance )

I need your help and support please
Please help
By what method are you introducing the gas mixture into the GC? And how are you handling from the gas cylinder to the instrument?

According to Agilent, the response factor for sulfur in independent of chemical composistion.http://www.chem.agilent.com/Library/tec ... 6785EN.pdf
I am working in speciality gases centre we are mixing gases with diffrent conc. As per custumer order .introduce the mixing gas to GC using gas regulator to regulate a gas pressure to GC
Thank you sir for your reply
I assume from a regulator through a gas sampling valve? I presume that you are using techniques which give good results with other gas mixtures. So, what surfaces is the gas mixture contacting? Sulfur compounds and surfaces such as metal can be a problem.
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry