Alternative SPE solvent?

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

4 posts Page 1 of 1
We're performing analysis of odor compounds in some fruit juices. One of the last steps of sample prep is elution of analytes from SPE column (Varian Mega B2 - C18) with dichloromethane. The question is which alternative solvent with higher boiling point might be used for washing the compounds from such a column?
Cantacoxinha wrote:
We're performing analysis of odor compounds in some fruit juices. One of the last steps of sample prep is elution of analytes from SPE column (Varian Mega B2 - C18) with dichloromethane. The question is which alternative solvent with higher boiling point might be used for washing the compounds from such a column?


It would depend on the compounds that you are looking for, but ethyl acetate is one that is used for environmental samples. It is more polar than dichloromethane, if you need similar polarity maybe hexane would work.
The past is there to guide us into the future, not to dwell in.
Thanks for replying. The compounds are thiols that are found in tropical fruits like passionfruit, guava etc. We are trying to substitute DCM with something less volatile since we're afraid that the analytes are partially lost when preconcentration of eluted solution from the SPE cartridge is done.
A suggestion. Rather than change the solvent used in eluting the SPE cartridge, add a "keeper" to the extract before evaporating away the DCM. A drop or two of a solvent like hexane, isooctane, or heavier. And then evaporate away the DCM, leaving the analytes of interest in the last remaining drops of solvent.
4 posts Page 1 of 1

Who is online

In total there are 4 users online :: 0 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry