Hi,

we performed a in vivo study with a radiopeptide, but the problem occured was, that if we use a gradient starting with 10% ACN-55% ACN 30min we get more peaks and one very high peak at about 4 min, whereas starting with 1% ACN - 55% ACN 30 min we get a lower amount of peaks and the very high peak at the beginning is missing.
Could anybody explain why we get so different results in the different gradient systems? An overlay of peaks was not observable... :cry:

greets Anton
Sounds like your detector is not selective enough (detects everything!). Since you are using a radiopeptide you should be using a scintillation detector or radioactive detector or MS (selective ion)
The high peak at 4 minutes when starting with 10% organic could be multiple analytes eluting together due to them not being as retained as they are when starting with 1% organic.

You need to inject a very high concentration of the labeled analyte in a clean matrix to establish the retention time for it in each gradient, then inject the sample and look for that peak at its proper retention time.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry