Diluents for VOC standards preparation

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

5 posts Page 1 of 1
What are the different diluents could be used for preparing VOC Standards?
Can methanol be used as the diluent for preparing the stock solution and water as the diluent for working standards (1,5,10,20 ppm)?
As long as methanol is not on your target list it is the preferred solvent for volatile organics analysis.
Unless you are doing an direct injection of an extract, your diluent will be water.
Yes, if doing purge and trap you will mostly be using methanol for the spiking solution then do dilutions of that into water for the standards you will purge. Check out EPA 8260 or EPA 624 for some good guidelines on standards for purge and trap volatiles analysis.

If not doing purge and trap, you usually use the same solvent your samples are prepared in.
The past is there to guide us into the future, not to dwell in.
Thanks for your suggestions!
Thanks for your suggestions!
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.