Good morning,

We're using EPA method 525.2 to confirm the absence of Benzo(a)Pyrene and Pentachlorophenol in drinking water. Both compounds are analyzed simultaneously in 1L samples. Sample prep is the following :

-Acidification to a pH lower than 2 using HCl 1N
-Addition of internal standards (Phenanthrene-D10 for PCP and Benzo(b)fluoranthen for B(a)P)
-Addition of surrogate standards (PCP-C13 for PCP and Benzo(e)pyrene for B(a)P)
-Addition of analytes of interest in laboratory fortified matrix samples
-Addition of 5mL Methanol
-Agitation for 1 minute

Compounds are extracted by SPE using C18 disks (3M 2215) :

-Disks are rinced using CH2Cl2
-Disks are activated using MeOH
-Sample (1L) is extracted
-Disks are dried 5 minutes
-Analytes are extrated using 3x5mL CH2Cl2
-Samples are dried using Na2SO4
-Samples are evaporated under nitrogen
-Samples are reconstituted using Toluene for analysis by GC-MS

In the 13 years I've been here, we've never once reported a result that wasn't lower than the reporting limit for a routine sample. The method is accredited and PT results have been satisfactory for many years now.

We're facing two recurring problems however :

1. Surrogate and matrix sample recovery for Pentachlorophenol is often above 130%
2. Benzo(a)Pyrene recovery in fortified matrix samples is often low (10-20%) while surrogate recovery is acceptable

Therefore, I have 3 questions :

1. What could account for the high recovery for Pentachlorophenol and should we be worried about this? Has anyone ever had an auditor question high recoveries when confirming the absence of analytes in a sample?

2. Is our low recovery of B(a)P merely matrix effect (since we don't have recovery problems for the surrogate) and if this is the case, do we have to take any action, or is the low recovery caused by sample preparation?

3. We've always added the internal standard during sample prep and are using a recovery standard added at the end of the analysis. Since surrogates standards are also used, has anyone tried placing the I.S. just before injection (thus replacing the recovery standard) without question by an auditor?

I would post a copy of our method but it's in French, so I apologize for the lack of details. Please note that we're situated in Canada, so we're not rigidly tied to the EPA method. Changes are acceptable if validation data is provided.

Thanks!
Mathieu
What you describe is close to 525.2, but ... Ethyl acetate is also used for the extraction. After the elution step the solvent is roughly 50:50 EA and MeCL2. For the concentration step the extract should not go less than 700ul.
I add my internals and surrogates into the conditioning methanol.
I have never had good luck with PCP using this method and do it via 515.
Prepare a blank with only solvent and the internal standards and compare it to the sample and see if the internal standard for PCP is having low recovery. If it is then that explains the high recovery for PCP and the Surrogate.

If you look at EPA 525.3 they use the labeled PCP as the internal standard for PCP which can solve such problems since they should both recover in a similar manner.

We are working on switching all of our drinking water semi volatiles over to 525.3. We found though that you have to use matrix matched standards for calibration. We don't extract the standards but instead extract blanks with nothing spiked into them, then use that to prepare the calibration standards so that you match the matrix with preservatives to account for any suppression or enhancement in the MS source. It works really well so far.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry