Hello. I know water is one solvent and it is highly soluble in water and easy to analyze on GC-FID. I have soil and water samples. I want to know, can I use dichloromethane to mix the water samples and separate them in separatory funnel. For the soil mix it with dichloromethane and then filter. Do you guys have any idea how to extract with solvent for both soil and water for GC-FID analysis. Thank you
-
- Posts: 3502
- Joined: Mon Aug 30, 2004 11:43 pm
You have a separate post about glycol assay. Do you mean ethylene glycol or one of the thousands of other gycols? I'm a smart guy, but I have no idea and don't want to wate my time on a guess.
I'd say to use methanol or dilute your aqueous extracts 1:1 with methanol.
Or - if you're going to derivatize to trimethylsilyl product - then use DMF or pyridine.
I'd say to use methanol or dilute your aqueous extracts 1:1 with methanol.
Or - if you're going to derivatize to trimethylsilyl product - then use DMF or pyridine.
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
Consumer Products Guy wrote:
You have a separate post about glycol assay. Do you mean ethylene glycol or one of the thousands of other gycols? I'm a smart guy, but I have no idea and don't want to wate my time on a guess.
I'd say to use methanol or dilute your aqueous extracts 1:1 with methanol.
Or - if you're going to derivatize to trimethylsilyl product - then use DMF or pyridine.
So i am analyzing ethylene glycol. I know water is one solvent i can use to mix glycol with. What about methanol. If we have water conaining glycol. Methanol is high soluable in water how can i seperate water from methanol and attach ethylene glycol to methanol. Thank you
-
- Posts: 3502
- Joined: Mon Aug 30, 2004 11:43 pm
You'd inject the water-methanol mix.
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
Consumer Products Guy wrote:
You'd inject the water-methanol mix.
Will that work?
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Direct inject the water + ethylene and propylene glycol onto a wax column with FID for detection. A 0.5 uL injection should see low ppm concentrations. For soils, extract with water, filter and analyze as a water sample.
It has been a while but this worked well for surface waters.
It has been a while but this worked well for surface waters.
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
Steve Reimer wrote:
Direct inject the water + ethylene and propylene glycol onto a wax column with FID for detection. A 0.5 uL injection should see low ppm concentrations. For soils, extract with water, filter and analyze as a water sample.
It has been a while but this worked well for surface waters.
i already know water works, but i want to compare with a solvent how does that do for extraction purposes
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
sohaib1 wrote:Consumer Products Guy wrote:
You have a separate post about glycol assay. Do you mean ethylene glycol or one of the thousands of other gycols? I'm a smart guy, but I have no idea and don't want to wate my time on a guess.
I'd say to use methanol or dilute your aqueous extracts 1:1 with methanol.
Or - if you're going to derivatize to trimethylsilyl product - then use DMF or pyridine.
So i am analyzing ethylene glycol. I know water is one solvent i can use to mix glycol with. What about methanol. If we have water conaining glycol. Methanol is high soluable in water how can i seperate water from methanol and attach ethylene glycol to methanol. Thank you
Essentially you can't separate the water from the methanol. The problem you will have with ethylene glycol is the hydrogen bonding. It is going to want to remain in the water more than any solvent that would not be soluble in water.
Acetonitrile can be separated from water by adding Sodium Chloride until you have a saturated solution, but then some will be dissolved in the Acetonitrile and give inlet problems after multiple injections.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
James_Ball wrote:sohaib1 wrote:Consumer Products Guy wrote:
You have a separate post about glycol assay. Do you mean ethylene glycol or one of the thousands of other gycols? I'm a smart guy, but I have no idea and don't want to wate my time on a guess.
I'd say to use methanol or dilute your aqueous extracts 1:1 with methanol.
Or - if you're going to derivatize to trimethylsilyl product - then use DMF or pyridine.
So i am analyzing ethylene glycol. I know water is one solvent i can use to mix glycol with. What about methanol. If we have water conaining glycol. Methanol is high soluable in water how can i seperate water from methanol and attach ethylene glycol to methanol. Thank you
Essentially you can't separate the water from the methanol. The problem you will have with ethylene glycol is the hydrogen bonding. It is going to want to remain in the water more than any solvent that would not be soluble in water.
Acetonitrile can be separated from water by adding Sodium Chloride until you have a saturated solution, but then some will be dissolved in the Acetonitrile and give inlet problems after multiple injections.
Thank you James very much for your help
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
- Location: UK
Hi Sohaib
Just to summarise
You are analysing for ethylene glycol in both soil and water samples.
For the soil
May I suggest that you just try the above suggestions using methanol or methanol /water and make your own comparisons and let us know how these compare with water only
Yes, but just try it for yourself ! and let us know the results . Methanol/water may work better.
Various soil types may have different adsorption properties for ethylene glycol
For the water
Yes, but you will have the partition co-efficient of the ethylene glycol between water and DCM to consider.
Easy for you to give it try and let us know the results
Why not just try adding some methanol ( 1:1 as previously suggested by CPG to reduce injection problems compared to just injecting the water sample) and filter?
In all cases above you will need to run spiked ( standard or multiple standard addition method ) and unspiked samples for your analysis to determine extraction efficiency.
Please let us know how you get on - it may help future posters with a similar problem
Just to summarise
You are analysing for ethylene glycol in both soil and water samples.
For the soil
i already know water works, but i want to compare with a solvent how does that do for extraction purposes .
May I suggest that you just try the above suggestions using methanol or methanol /water and make your own comparisons and let us know how these compare with water only
For the soil mix it with dichloromethane and then filter.
Yes, but just try it for yourself ! and let us know the results . Methanol/water may work better.
Various soil types may have different adsorption properties for ethylene glycol
For the water
I want to know, can I use dichloromethane to mix the water samples and separate them in separatory funnel.
Yes, but you will have the partition co-efficient of the ethylene glycol between water and DCM to consider.
Easy for you to give it try and let us know the results
Why not just try adding some methanol ( 1:1 as previously suggested by CPG to reduce injection problems compared to just injecting the water sample) and filter?
In all cases above you will need to run spiked ( standard or multiple standard addition method ) and unspiked samples for your analysis to determine extraction efficiency.
Please let us know how you get on - it may help future posters with a similar problem
Regards
Ralph
Ralph
-
- Posts: 56
- Joined: Tue Feb 06, 2018 10:38 pm
GOM wrote:
Hi Sohaib
Just to summarise
You are analysing for ethylene glycol in both soil and water samples.
For the soili already know water works, but i want to compare with a solvent how does that do for extraction purposes .
May I suggest that you just try the above suggestions using methanol or methanol /water and make your own comparisons and let us know how these compare with water onlyFor the soil mix it with dichloromethane and then filter.
Yes, but just try it for yourself ! and let us know the results . Methanol/water may work better.
Various soil types may have different adsorption properties for ethylene glycol
For the waterI want to know, can I use dichloromethane to mix the water samples and separate them in separatory funnel.
Yes, but you will have the partition co-efficient of the ethylene glycol between water and DCM to consider.
Easy for you to give it try and let us know the results
Why not just try adding some methanol ( 1:1 as previously suggested by CPG to reduce injection problems compared to just injecting the water sample) and filter?
In all cases above you will need to run spiked ( standard or multiple standard addition method ) and unspiked samples for your analysis to determine extraction efficiency.
Please let us know how you get on - it may help future posters with a similar problem
Methanol didn't work at all. Only water worked so for soil i did 1 Soil and 2 water ratio. Methanol didn't work all i go was big methanol peak nothing else
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
- Location: UK
Thank you for the update, feedback and for giving the suggestions a try.
An interesting but slightly perplexing result since I would expect EG to be completely miscible with methanol - please correct me if I am wrong - I need to give it some thought
An interesting but slightly perplexing result since I would expect EG to be completely miscible with methanol - please correct me if I am wrong - I need to give it some thought
Regards
Ralph
Ralph
-
- Posts: 44
- Joined: Tue Oct 04, 2011 9:14 am
Admittedly a different matrix (alcoholic beverages) and a different detection technique (MS), but we use ethyl acetate as the solvent. The down side is you need to salt out the aqueous samples.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
GOM wrote:
Thank you for the update, feedback and for giving the suggestions a try.
An interesting but slightly perplexing result since I would expect EG to be completely miscible with methanol - please correct me if I am wrong - I need to give it some thought
I would imagine that since ethylene glycol is not retained that strongly it may be lost in the large methanol solvent peak. One reason aqueous injections work better for this on FID.
The past is there to guide us into the future, not to dwell in.
14 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science