Can anybody suggest how to prepare QC samples. I am analyzing Ethylene and Propylene Glycol. Can i prepare QC from Stock Calibration solution or it has to be diffrent. I need help in QC Please. Thank you in advance
It depends on what agency you are reporting data to.

Most require that a calibration curve be validated by a QC sample that is prepared from a source separate from that of the calibration curve standards, but ongoing QC can be from either source. If you do not have two sources of ethylene and propylene glycols, then make two separate preparations of a stock dilution standard from the original compounds. This will show at least that there is no problems with the preparation.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
It depends on what agency you are reporting data to.

Most require that a calibration curve be validated by a QC sample that is prepared from a source separate from that of the calibration curve standards, but ongoing QC can be from either source. If you do not have two sources of ethylene and propylene glycols, then make two separate preparations of a stock dilution standard from the original compounds. This will show at least that there is no problems with the preparation.


So are you suggesting i make separate QC from pure Ethylene Glycol and Propylene Glycol. Like lets say make 50 PPM Ethylene Glycol and 50 PPM Propylene Glycol?
Why not just obtain a previously tested QC sample and spike it with various concentrations of Glycol and EO?
sohaib1 wrote:
James_Ball wrote:
It depends on what agency you are reporting data to.

Most require that a calibration curve be validated by a QC sample that is prepared from a source separate from that of the calibration curve standards, but ongoing QC can be from either source. If you do not have two sources of ethylene and propylene glycols, then make two separate preparations of a stock dilution standard from the original compounds. This will show at least that there is no problems with the preparation.


So are you suggesting i make separate QC from pure Ethylene Glycol and Propylene Glycol. Like lets say make 50 PPM Ethylene Glycol and 50 PPM Propylene Glycol?


If you have a pre-made mixture from a supplier then you can use that also. We normally order a solution from one supplier to use for making the calibration curve and another solution from a different supplier to do a QC standard. Restek, AccuStandard, Ultra, or any of the many other suppliers will probably have what you need for about $35.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
sohaib1 wrote:
James_Ball wrote:
It depends on what agency you are reporting data to.

Most require that a calibration curve be validated by a QC sample that is prepared from a source separate from that of the calibration curve standards, but ongoing QC can be from either source. If you do not have two sources of ethylene and propylene glycols, then make two separate preparations of a stock dilution standard from the original compounds. This will show at least that there is no problems with the preparation.


So are you suggesting i make separate QC from pure Ethylene Glycol and Propylene Glycol. Like lets say make 50 PPM Ethylene Glycol and 50 PPM Propylene Glycol?


If you have a pre-made mixture from a supplier then you can use that also. We normally order a solution from one supplier to use for making the calibration curve and another solution from a different supplier to do a QC standard. Restek, AccuStandard, Ultra, or any of the many other suppliers will probably have what you need for about $35.


OH ok thank you. So what QC will i need to make. The results will be reported to an internal department.
6 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry