Backpressure Regulator Introducing Pulsating Flow

Posted: Thu Jun 15, 2017 10:58 pm
by thejsychemist
Dear all,

In our lab we use a constant volume Eldex Optos 2SI single-head reciprocating pump (PEEK flow path; ruby seats and sapphire balls) to create a very thin liquid jet for spectroscopic study of the resultant jet solution. Since this single-head HPLC solvent pump introduces pulsing (which is terrible if you want to overlap the jet with a laser, for example) two pulse dampeners downstream of the HPLC pump were introduced. (SSI Lo-Pulse LP-21; PEEK). Inlet tubing is 1/8'' OD PTFE and the outlet tubing is 1/16'' OD PEEK rated to 5000 psi.

These pulse dampeners dampened the pulsations considerably. Since these dampeners are rated to work optimally at 1000 psi, I introduced a 1000 psi backpressure regulator (Upchurch; PEEK) after both the pulse dampeners, however the pulsations again returned after installation of the regulator.

I figure that I must be doing something fundamentally wrong with the backpressure regulator such that I re-introduce pulsations. Is it a matter of tightening the backpressure assembly at a given flow-rate such that the pulsations disappear? I'm unsure as to whether the regulator needs recalibration at different flow rates.

All the best,
Ryan

Re: Backpressure Regulator Introducing Pulsating Flow

Posted: Mon Jun 19, 2017 2:39 am
by tom jupille
You said that the pulse dampeners reduced the pulsation "considerably". Was the level acceptable? If it was, I would advise you "if it ain't broke, don't fix it!"

If it's still not scceptsble, then yes, you should fine-tune the regulator.

Re: Backpressure Regulator Introducing Pulsating Flow

Posted: Mon Jun 19, 2017 5:29 am
by thejsychemist
Dear Tom,

Thanks for your reply. The maximum flow rate for the pump is 10 ml per min. Installing two pulse dampeners made it possible to achieve a flow rate of about 5 ml per minute without any appreciable pulsing.

I was of the opinion that since the pulse dampeners work optimally with a backpressure of 1000 psi then installing a backpressure regulator would mean that I could achieve a lower flow rate (1 ml per minute or less) without pulsing; this appears to be wishful thinking.

A lower flow rate as possible is better for this application (since I don't have to re-fill my reservoir of solution, which is critical because the solution is air and moisture sensitive). Adjusting the tightness of the backpressure regulator does not seem to bring about a change in the pulsing.

All the best,
Ryan

Re: Backpressure Regulator Introducing Pulsating Flow

Posted: Mon Jun 19, 2017 1:49 pm
by Steve Reimer
If you want a lower flow but need to be at 10 mL/min for optimum pulse damping, would it be practical to split the flow with a return line to the reservoir?

Re: Backpressure Regulator Introducing Pulsating Flow

Posted: Tue Jun 20, 2017 8:42 pm
by James_Ball
Steve Reimer wrote:
If you want a lower flow but need to be at 10 mL/min for optimum pulse damping, would it be practical to split the flow with a return line to the reservoir?


This is what I was thinking also. It might be possible to configure it so that the pulses travel in the return line to give a more steady flow to the end point of the system. Possibly using the packpressure regulator after the split so that what is coming out at the end has a steady pressure.

Re: Backpressure Regulator Introducing Pulsating Flow

Posted: Fri Jun 23, 2017 3:47 am
by thejsychemist
Dear all,

Yes that might be a sensible option.

Thanks for all of your help everyone.

All the best,
Ryan