amlodipine besylate adsorption on filter

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

3 posts Page 1 of 1
adsorption of amlodipine besylate on filter after dissolution ask if any another method to exclude filtration after dissolution with put onto consideration centrifuge perform the same problem on dissolution results can i take sample as it is for injection on HPLC ???
formulator
Several comments here:
1. Have you tried different filter media to find something less adsorptive?
2. Centrifugation should work. However if you find evidence of adsorption on the pellet (i.e. incomplete recovery from a spiked matrix blank), then you really have to revisit your formulation, because that suggests that you will get poor availability in vivo
3. Direct injection into the HPLC may be possible if the volume injected is small enough, but I would *definitely* use an in-line filter and/or a guard cartridge -- and be prepared to change them frequently.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
centrifuge perform the same problem can i take sample as it is from dissolution vessel to inject it to HPLC direct ???
formulator
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry