- Posts: 69
- Joined: Mon Jul 11, 2011 3:15 pm
- Location: Somerset, PA, USA
EPA 552.3 uses the ISTD in the extraction solvent, we have found it works quite well. I have less extraction variablity and better recovery, probably due to using more acid at derivatization and no CuSO4. I also use a different surrogate, the method surrogate has given me wildly different results between batches/technicians.
-
- Posts: 8
- Joined: Thu Oct 18, 2012 9:46 pm
- Location: NJ, USA
Have any of you had any problems with interferences in the chromatograms when doing 552.3? I'm seeing a large number of interference peaks all over my chromatogram especially near the surrogate on my primary column I have very consistent recoveries on one column whereas on the column the results seem to be all over the place.
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science