We are using liquid-liquid extraction method with extraction solvent ethyl acetate. We have preformed recovery testing by comparing area ratios of analyte and IS in IS post-spiked (IS post spiked after QC extraction) samples with pure analytical solution. We got 17.4% and 20.4% CV for analyte 1 and analyte 2.
All other tests (Pr and Acc, stability etc.) passed.
We set 15% acceptance criteria on %CV.
According to FDA, recovery should not be 100%, but it should be precise at each level.
Even if the recovery is not precise, because LLE is a technique with high variability, IS can deviate from extraction to extraction: do we need to have criteria on recovery CV, if we are using SIL IS which will compensate for any deviation during extraction?
Could you say how you are doing recovery test?